Hi,
We recently changed form Chemstation to OpenLab CDS and I tried to introduce it on the report of ISTD method but the result was 0.00 or #Error. What would be the correct formula?
Find maintenance, best practice, and troubleshooting articles.
Hi,
We recently changed form Chemstation to OpenLab CDS and I tried to introduce it on the report of ISTD method but the result was 0.00 or #Error. What would be the correct formula?
Hi Andy,
The two formulas that we tried was:
=Sample_Amount/Peak_Area --> 0.00
=Sample_DAInternatalStandardAmounts --> #Error
The version of OpenLab CDS is 2.5
Elena
Elena,
I am not sure exactly how the value is helpful but Compound_Amount/Peak_Area is likely what you are looking for from the post question.
Marty Adams
Perfect! it was the solution! Thank you!
HI Elena
Sample_Amount is not a calculated value, but the number typed in the Sample amount column in the sequence table. As martin.adams says, the correct field is Compound_Amount for these calculations.
/Andy