Hello,

I have a problem with launching Acquisition module while repeatedly error occurs: "Internal server error".

After trying to launch Instrument following description of error occurs..

I tried to restore windows to the time it was working and restart all "agilent" services, bud it didnt work. BTW data analysis etc. works... Could you help me with this?

Thank you

Parents
  • I wanted to "repair" OpenLab in original  installer file, but there is no, only uninstall button (see picture below)... If i uninstall it and then install it back, will it cost me generating new licence key, or old licence key will work? Thank you

  • I solved same error yesterday on customer site. DB is corrupted probably, so uninstall/install is not enough as the corrupted DB will be used again. You should uninstall OpenLAB and purge (Shift-Delete) folders C:\ProgramData\Agilent, C:\program Files(x86)\Agilent Technologies as well. Database will be erase.

    Than you can install OpenLAB again and create instruments, create same projects as you have in C:\CDSProjects\ (if you use default). Make sure to backup your data, for safe.

    Majkee

Reply
  • I solved same error yesterday on customer site. DB is corrupted probably, so uninstall/install is not enough as the corrupted DB will be used again. You should uninstall OpenLAB and purge (Shift-Delete) folders C:\ProgramData\Agilent, C:\program Files(x86)\Agilent Technologies as well. Database will be erase.

    Than you can install OpenLAB again and create instruments, create same projects as you have in C:\CDSProjects\ (if you use default). Make sure to backup your data, for safe.

    Majkee