We have an autosampler (G1313A) on our 1100 HPLC with tray chiller installed; however, there is not option in OpenLab to set the tray temperature. We are using OpenLab server 3.2.0. What are we missing?
We have an autosampler (G1313A) on our 1100 HPLC with tray chiller installed; however, there is not option in OpenLab to set the tray temperature. We are using OpenLab server 3.2.0. What are we missing?
Hi sassman,
When selecting Thermostat controlled by method, you should be able to change it in every method when loading it in the software.
When not selecting it, you could right click on the instrument -> control -> and you should see an option to change temperature.
Also, you can easily modify this setting by going to right click -> Modify -> temperature setting, and again it gives you the option to choose between Instrument controlled or method controlled.
Thank You! I didn't know about Modify > temperature setting. That's exactly what we needed.