Hi, I want to report the number of injections made in a sequence without using a custom calculation but can't find a field to use for this. Is there one available in OpenLab 2.3 please?
Kind Regards,
Camelia
Find maintenance, best practice, and troubleshooting articles.
Hi, I want to report the number of injections made in a sequence without using a custom calculation but can't find a field to use for this. Is there one available in OpenLab 2.3 please?
Kind Regards,
Camelia
Hello,
I have used =CountDistinct(Injection_AcquiredDate) or =CountDistinct(Injection_ID) as a field on a report. If you are doing a cross sequence summary then group and repeat on sequence name or ID.
Marty,
Thank you (once again) for your help. I have used the syntax and now created a cross sequence summary report to output the number of injections per sequence, the total number of injections across the sequences alongside the number of sequences. This could be useful in the future.
Kind Regards
Marty,
Thank you (once again) for your help. I have used the syntax and now created a cross sequence summary report to output the number of injections per sequence, the total number of injections across the sequences alongside the number of sequences. This could be useful in the future.
Kind Regards