Haw can I calculate the average of peak area for alternating injections
Find maintenance, best practice, and troubleshooting articles.
Hello
I have a question
Haw can I calculate the average of peak area for alternating injections
Hello,
In order to assist you we will need the following:
I use openlab CDs
Hello,
Is there something in common between the injections that you want to average(sample name, vial position, label, ect.) Below is a CC example that averages peak areas for identified peaks for all injections in the same vial.
CurrentSequence.AllIdentifiedPeaks.Where(function (x) x.Compound_Name = CurrentPeakOrGroup.Compound_Name).Where(function (x) x.Injection.Sample.Sample_VialNumber = CurrentInjection.Sample.Sample_VialNumber).Select(function(x) x.Peak_Area).Average
Marty Adams
Hi Rahma05
Do you need to do this on an automated basis. I use the Chemstation version of Open Lab and if you select any two or more results in data analysis and the appropriate report layout you can get this information displayed.
Not sure if this helps
Kevin
Hello,
This CC would average the current injection with the next by Sample_OrderNumber in the result set. This would be a more generic way to get the average.
CurrentSequence.AllIdentifiedPeaks.Where(function (x) x.Compound_Name = CurrentPeakOrGroup.Compound_Name).Where(function(x) x.Injection.Sample.Sample_OrderNumber = CurrentSample.Sample_OrderNumber Or x.Injection.Sample.Sample_OrderNumber = (CurrentSample.Sample_OrderNumber +1)).Select(function(x) x.Peak_Area).Average