I can see the entire run while I am using Chemstation online, but when I go to integrate the scan, only a half to a third is shown while in Chemstation offline.
Find maintenance, best practice, and troubleshooting articles.
I can see the entire run while I am using Chemstation online, but when I go to integrate the scan, only a half to a third is shown while in Chemstation offline.
What revision of ChemStation are you running?
In Offline Data Analysis if you go to Graphics>Signal Options do you have a custom range specified?
What revision of ChemStation are you running?
In Offline Data Analysis if you go to Graphics>Signal Options do you have a custom range specified?
I am using Rev.B.04.03-SPI{87}, and I have the range set to full. Just tried to change it to 0-16 which should show the whole scan, but the same issue occurs.
In that case then this could indicate that the conf_off.reg file for the instrument is corrupt. Try deleting the conf_off.reg file located at C:\Chem32\#\ where # is the instrument number. Then launch an offline session and see if the problem is still there.