I'm having issue generating a report template that is capable of telling me the resolution between 2 peaks, can someone assist me with entering the formula for this in OpenLab 2.2?
I'm having issue generating a report template that is capable of telling me the resolution between 2 peaks, can someone assist me with entering the formula for this in OpenLab 2.2?
Hello,
I marked your post as a question and added tags to increase visibility.
OpenLab CDS already has the built in functionality to calculate peak resolution so there is no need to manually enter the formula into a custom calculation file or report template.
In order to calculate peak resolution, all you need to do is update the System Suitability parameters in your Processing Method to evaluate column performance for all peaks and specify a pharmacopeia (EP, USP, or JP):
The resolution can then be called on the report using the value =Peak_Resolution_USP (or Peak_Resolution_EP, etc...).