Hey guys, I'm getting this error when trying to log into Chemstation for our 1200 HPLC: Configuration system failed to initialize. Chemstation will shutdown now. ANY help would be appreciated!
Find maintenance, best practice, and troubleshooting articles.
Hey guys, I'm getting this error when trying to log into Chemstation for our 1200 HPLC: Configuration system failed to initialize. Chemstation will shutdown now. ANY help would be appreciated!
Hi Noemidelgado,
The files should be recreated. Do you know if you are hiding the file extensions? In other words are you seeing, for example, config or config.reg in that folder? The file .old extension must be added after the file extension (config.reg.old).
Hello,
Do you have more than one instrument configured in the OpenLab Control Panel? If so, make sure that you are renaming the files for the correct instrument. Changes made to the configuration files in C:\Chem32\1 will only impact instrument 1, C:\Chem32\2 only impacts instrument 2, etc...