Post-run Macro to export Signal Raw Data to ASCII

Dear All,

 

I'm using a GC system 7890A and controlling it with Chem-Station software.

 

I was wondering if anyone can help me in finding a way to automatically export the signal raw data to an ASCII file, ideally also with the option of choosing the folder destination.

 

I have already tried to modify the report output, however I managed to produce only a .txt file containing the integrated peaks' area, but not the signal raw data (X and Y values).

 

I think the answer lies in writing a post-run command/macro to be inserted when I edit the method that I use, however I am really unfamiliar with macro running and would be grateful if you could help me in this, or any other solutions I am not aware of.

 

Thanks in advance for your time

 

Best Regards,

 

Amedeo

Parents
  • Hi Amedeo,

    You should be able to do this with a simple postrun macro.  The signal data is suppose to be stored in the ChromReg register.  The "Macro programming guide" covers the basics of this.  You just need to export it out of the register.  I believe Agilents website still has a downloadable version of this old but applicable manual.

    If you don't want to try programming this yourself,  I'll try and write something for you when I have some spare time.

    Tim

Reply
  • Hi Amedeo,

    You should be able to do this with a simple postrun macro.  The signal data is suppose to be stored in the ChromReg register.  The "Macro programming guide" covers the basics of this.  You just need to export it out of the register.  I believe Agilents website still has a downloadable version of this old but applicable manual.

    If you don't want to try programming this yourself,  I'll try and write something for you when I have some spare time.

    Tim