reporting pooled results with unknown number of samples

HI there,

I'm looking to report pooled precision from sequences but do not know how many samples wil be run.

The sequence takes the format:

inj/vial
Blank1
CAL 1-61
Precision (Cal 3)6
Sample 1-51
Precision 2 (Cal 3)1
Sample 6-101
Precision 3 (Cal 3)1
Sample 10-131
Final Precision 4 (Cal 3)1

 

The first set of precision data uses 6 replicate injections then after every 5 samples there is a single injection of the same cal standard and the data is combined with the Precision data from the 6 injections and the pooled RSD is then also reported. The problem is I don't know how many samples ie. how many individual precision injections (after every 5 samples) will take place in the sequence and can't think how to set the report up to take this into account.

If you are able to suggest a way to approach this it would be much appreciated.

Many thanks

Parents
  • Hi Marty,

    Thanks for your reply and suggested CC approach. I will try to clarify what i want the report to do. i have attached an excel file showing an example but it doesn't open properly for me so i have inserted the table below.

    I want to report the pooled RSD for Precision 2 (1injection) + Precision (6 injections) = count 7; then report separately the pooled RSD for Precision 3 (1 injection) + Precison 2 (injection) + Precision (6 injections) = count 8. And so on for each check run every 5 samples. So if the pooled RSD for Precision 3 failed (injected after samples 6-10) but Precision 2 passed (injected after samples 1-5), samples 1-5 would be valid but samples 6-10 would not be valid. If there were a fixed number of samples run each time i could do it but  i don't know the number of samples going to be run and therefore i dont know the number of these Precision (single injections that need to be pooled with the preceding precision data (6 injections + single injection of each subsequent precision injection). Apologies for struggling to explain this rather odd requirement. I look forward to any further comment you may have. 

     

    Area Compound A 
    6 replicate Precision injections20
    21
    20
    16
    15
    20
    Mean18.66666667
    SD2.503331114
    RSD 6 replicate Precision injections13%count = 6
    Precision 222
    Mean19.14285714
    SD2.60950643
    RSD Pooled Precision 2 14%count = 7
    Precision 325
    Mean19.875
    SD3.181980515
    RSD Pooled Precision 316%count = 8
  • In this scenario is the next bracket also a count of 8? Would the results be the first 6, 3 and 4? If so you could do something like I have below. The first CC creates an index of all runs with the sample name CCC. Then the second CC references the first and uses that index plus a filter on sample type cal standard to do the calculation. You can change the filtering to other fields as needed. If you want to roll up all of the bracket injections up to the current one, you could change the query expression to be Where(function x.Injection.GetIntegerCC("CSSIndex") <= CurrentInjection.GetIntegerCC("CSSIndex") OR x.Injection.Sample.Sample_Name = "SSS". I have attached a notepad file with the expressions shown because the editor is displaying "(x)" as a .

     

    Scope: Injection Type: Integer

    CurrentSequence.AllInjections.Where(function x.Sample.Sample_Name = "CSS").OrderBy(function x.Sample.Sample_OrderNumber).Select(function x.Sample.Sample_OrderNumber).ToList.FindIndex(function x=Sample_OrderNumber)+1

     

    Scope: Peak or Group Type: Double

    RelStdDev("Peak_Area",CurrentSequence.AllIdentifiedPeaks.Where(function x.Injection.Sample.Sample_Type = 1).Where(function x.Injection.GetIntegerCC("CSSIndex") = CurrentInjection.GetIntegerCC("CSSIndex") OR x.Injection.Sample.Sample_Name = "SSS" OR x.Injection.GetIntegerCC("CSSIndex") = CurrentInjection.GetIntegerCC("CSSIndex")-1).Where(function x.Compound_Name = CurrentPeakOrGroup.Compound_Name))

     

     

     

    CCF.txt.zip
  • Hi Marty

    The next bracket would be 9 - each subsequent check would increase by a count of 1 as it would be pooling that single injection with the previous SS checks. Thanks for attaching the CC's in notepad. It will take a while for me to test this out as I'm working from home at the minute and unfortunately don't have access to the software. I have attached an example showing where in the run I want to report the RSD results. If you have any further comments on this please let me know.  Thanks very much for your input I will get back to you when I can.

    Cheers

    count eg.txt.zip
  • Okay if that is the way you are doing it then you can forget custom calculator although that would work with the <= index. The way you want to do this can be done in IR. The example below shows all the standards in one repeating table. The report shows 2 ways to present the values in the table or as fields below. The template relies on a filter of sample type as calibration standards but that could be changed to something else. It also assumes that the sample names of the first injection set are all the same and that the other standards have unique names. You could alter this to use another method to repeat them as needed. This method stores the peak_area as an aggregator indexed on compound_name in case you have more than on compound.  I am attaching the example template so you can see the way I did it. The template was done in 2.4.

     

     

     

    RSD Testing.rdl.zip
Reply
  • Okay if that is the way you are doing it then you can forget custom calculator although that would work with the <= index. The way you want to do this can be done in IR. The example below shows all the standards in one repeating table. The report shows 2 ways to present the values in the table or as fields below. The template relies on a filter of sample type as calibration standards but that could be changed to something else. It also assumes that the sample names of the first injection set are all the same and that the other standards have unique names. You could alter this to use another method to repeat them as needed. This method stores the peak_area as an aggregator indexed on compound_name in case you have more than on compound.  I am attaching the example template so you can see the way I did it. The template was done in 2.4.

     

     

     