Lock a master acquisition method in a project -

Hello, 

We are starting to use OpenLAB 2.3. The way to quantify for some products is the same.  How can we introduce an intelligent report within each of the projects without having to create it from 0 in each of them?

Regarding the acquisition method, as a supervisor can leave a method blocked so that the analyst/technician at the time of launching the sequence can not change the injection parameters (flow, volume,...). The option that we have been told from a near company is to create a "ghost" folder where the method is created, it is saved and locked in the ECM by administrator and then an administrator transfers it to the project folder. Is this true? 

Thanks a lot!

Òscar 

Parents
  • Former Member
    +1 Former Member

    Hi Oscar

    As Marty said there is option to share Report templates across the projects by creating Generic project for Report templates.

    Generic projects can be created after unchecking 'OpenLab CDS' option from project creation window.

    If you don't want user to edit Acquisition method then simply remove 'Create and modify acquisition method' privilege from user roles under Project.

    Sidharth

Reply
  • Former Member
    +1 Former Member

    Hi Oscar

    As Marty said there is option to share Report templates across the projects by creating Generic project for Report templates.

    Generic projects can be created after unchecking 'OpenLab CDS' option from project creation window.

    If you don't want user to edit Acquisition method then simply remove 'Create and modify acquisition method' privilege from user roles under Project.

    Sidharth