Hi community, please help me, I can use ferrule part number 5180-4109 with this trap MT400-2 or not.
Please, recommend me ferrule for the Moisture trap.
Hi community, please help me, I can use ferrule part number 5180-4109 with this trap MT400-2 or not.
Please, recommend me ferrule for the Moisture trap.
Those are swagelok ferrules and they should be fine on 1/8th inch gas connection for that trap.
Regards
James
Ferrules 5180-4109 suit for the trap MT400-2?
This ferrules has other size that ferrules from traps kit.
I attach a photo.
Ferrules 5180-4109 suit for the trap MT400-2?
This ferrules has other size that ferrules from traps kit.
I attach a photo.
james_jenkins , thanks!
I have used both types without issue, usually , if I had extra swagelok fittings, I prefer to use them rather than the supplied ferrules.
Regards
James
Hi a.fedulova,
Ferrules 5180-4109 suit for the trap MT400-2!
MT400-2 has 2 ferrules include. It is compatible with tube 1/8 inch. You can buy nut and ferrule as a part backup of this installation
+ 5080-8750 1-8 Brass Nut and Ferrule Set ( front and back), 20pk Qty:1
or + 5080-8751 Nut and ferrule set, 1-8 in, stainless steel, 20 pk Qty:1
Sincerely
John