It is possible to use the flowmeter after the cartridge calibration is expired without changing the cartridge?
It is possible to use the flowmeter after the cartridge calibration is expired without changing the cartridge?
Hi James,
thanks a lot!
Guillermina
Hi guilleurreta,
The GC forum doesnt support the flow meter , please contact GC-column-support@agilent.com and they should be able to provide support. I did move your post to our consumables forum and they may be able to provide the needed support.
James
Thank you for your answer Don!
Yes. The cartridge remains operational after the cartridge calibration has expired. You will continue to get the "expired" message and the measurement will not be traceable since it is out of calibration, but it will continue to function and measure flow.
Don Gage
CSD Technical Support Channel Manager