This Information Applies To: All Agilent GC systems with flame ionization detector (FID)
Issue:
The FID uses a hydrogen flame to burn organic samples, resulting in increased ionization that is measured as a current from 5 to 10^6 pA. Normal FID...
This Information Applies To: Agilent GC Systems with Split/Splitless Inlet.
Issue
A leak may produce the following symptoms:
Inlet pressure unable to reach setpoint (probably a large leak) - will lead to a flow shut-down.
Failure of Any Diagnostic...
This Information Applies To: Agilent OpenLab C DS software (Rev. 2x)
Issue
If the column configuration has been defined in the method, the instrument will calculate the pressure to maintain the target column flow rate. The column configuration must...
This Information Applies To: Agilent 7697A Headspace Sampler
Issue
The transfer line will need to be replaced if it breaks at the valve, the inlet or in the middle of the transfer line. The previously broken liner will need to be removed and new...
This Information Applies To: Agilent GC Systems 6890N, 7890A/B, 7820, 8860,8890, Intuvo 9000
Issue
The communication between Agilent GC instrument and the computer (PC) requires an active LAN connectivity. Each part of the instrument-PC network...