Using GC / MS - single quad, with extraction in SPE cartridge.
Splitless injection mode
HP-5MS UI column (30 m x 0.25 mm x 0.25 µm)
Using GC / MS - single quad, with extraction in SPE cartridge.
Splitless injection mode
HP-5MS UI column (30 m x 0.25 mm x 0.25 µm)
I typically do this was a pulsed splitless injection. 50 psi pulse for .5 min and purge flow to split vent of 50ml/min at .48 min. I typically use a single tapered direct connect liner in the GC inlet.