does anyone have/develop a method for method/CO2 analysis using HP-PLOT U column, TCD and helium as a carrier gas?
does anyone have/develop a method for method/CO2 analysis using HP-PLOT U column, TCD and helium as a carrier gas?
Thank you for your answer. What temperatures of the injection port and TCD would you recommend? I have seen 150C for injection, 250C for TCD and split ratio of 10-30:1 for other analysis methods but am not sure for my case.
Hi,
Here's few links that may help you out. In first presentation there's reference to Application Note 228-387 where they have used PLOT Q.
https://www.agilent.com/cs/library/applications/HPI_Compendium.pdf
https://www.agilent.com/cs/library/eseminars/Public/PLOT%20Column%20Selection.pdf
Hi khemmathini,
I just wanted to follow-up to see if this issue is resolved. If so, please come back to the post and click the “Correct Answer” button on the response that help so it will make the solution more visible. If you still need help, just let us know and we would be happy to continue working with you. I also moved this to applications energy and chemicals for better visibility.
Hi khemmathini,
Please provide your system details also so according solution will provide.
If in your sample CO2 is greater then 20 ppm then without methanizer you can got the response or if want lower detection then you want methanizer.
Oven- 40 C ----- 6 min then 20 C upto 200C final hold 1 min.