This Information Applies To: Agilent OpenLab CDS
Issue
This article outlines a basic workflow from loading raw data to generating a report in OpenLab CDS Data Analysis.
Steps to Follow
Load data
Analyze or process data
Review data
Generate report
Load Data
Launch Data Analysis...
This Information Applies To: Agilent OpenLab ChemStation
Issue
If the DAD full spectrum was saved during the run, it is possible to extract a signal at a different wavelength from the originally acquired in the acquisition method. This article describes how to extract chromatograms at specific...
This Information Applies To: Agilent OpenLab CDS, OpenLab ChemStation Edition C.01.xx, G4260A 380 ELSD, G4261A 385 ELSD, G4260B 1260 Infinity ELSD, G4261B 1290 Infinity ELSD, and G7102A 1290 Infinity II ELSD.
Issue
Outlining step-by-step instructions how to configure or reconfigure an Agilent evaporative...
This Information Applies To: Agilent OpenLab CDS 2.x in file system configuration
Issue
Sometimes, you will need to change the Projects Root Path / Project Storage Path . This article outlines the procedure on how to change it.
Background
For Standalone OpenLab CDS Workstation, all project...
This Information Applies To: Agilent OpenLab CDS 2.x
Issue
This article outlines the step-by-step instruction on how to overlay chromatograms for OpenLab CDS 2.x
Steps to Follow
Load data from result sets or single samples. See Load or add data .
Select the chromatograms that you want...