Calculate RSD on alternating injections

Hello,

 

I want the be able to calculate the RSD of our reference injections, but we need to do it each time with a different injection. I'm out of inspiration to get them grouped together. This is the situation:

 

1) at the beginning of the sequence, the reference is injected 5 times

2) after 4 sample injections the reference is repeated

3) in total this are 6 reference injections, RSD needs to be calculated on those

 

So far no problem, but we can build the sequence endlessly with sample injections. So if we have 8 sample injections, another reference injection is made. Then we need to calculate RSD of this reference injection with the 5 injections at the beginning, but without the reference injection made after 4 sample injection.

 

And so on for more sample injections. I don't know how to group every time the first 5 reference injections, with each other reference injection sepperately.

 

Thanks in advance,

Davy

  • Hello,

     

    I'm the analyst trying to calculate the RSD on the control reference in our sequence.

     

    I created a ccf file and made the link with the processing method used in the result set. Yet, no additional result is seen in the injection results after reprocessing all injections from the result set.

     

     

    First: do I have to change something in my injection list? The sample label field is now empty, what do I need to insert here?

     

    Second: how do I continue? = how do I import the results from this ccf result into the report editor? How do we filter in the injections using the ccf?

     

    Thanks for your input!

     

    Kevin

  • Hello,

     

    You will need to fix the syntax of the CC as it looks like you have an extra 1 at the end. There is a check file button in the CC editor that can be used to check all expressions in the file. In the sample label field,  I used SSS to isolate the initial standard group.

     

    Marty Adams 

  • Hello,

     

    Thanks for the update, I've corrected my typo.

     

    Yet, receive a lot of errors when applying the "check file button". (see attachment)

    Could you tell us what to replace the "x" with?

     

    And what does the "SSS" in the formula stand for?

     

    We're all new in the custom calculations part of the software...

     

    Thanks

     

    Kevin

  • Kevin,

     

    Sorry, I missed this in the earlier reply but the editor in the community page removed all the (X) values from the syntax of the CC. Try the syntax below it should work as I just tested it again. In my example, the SSS is the value I placed in the sample label field for the initial block of standards. 

     

     

    RelStdDev("Peak_Area",CurrentSequence.AllIdentifiedPeaks.Where(function (x) x.Injection.Sample.Sample_CalibrationLevel = 1).Where(function (x) x.Injection.Injection_SampleLabel.ToUpper.Equals("SSS") OR x.Injection.CreationTime = CurrentInjection.CreationTime).Where(function (x) x.Compound_Name = CurrentPeakOrGroup.Compound_Name))
  • Hello Martin

    sorry to necro this post :s

    I just wanted to say that we have recently had some training about CC's, and one of the things we wanted to implement was this feature.
    We used your formula, (our instructor said it was quite original with the "OR" in it and quite elegant) and have successfully used it!!

    A year ago our knowledge was not quite there yet, and my colleague Kevin wasn't able to make it work.
    Now we actually understand what they do Slight smile

    so (a late) thank you for this.

    Kind regards

    Sven

Was this helpful?