This Information Applies To: 5977A and 5977B
Content: This procedure is to guide the installation of a capillary column on 5977A/B GC/MS with a standard column nut or self-tightening nut.
Warning : a. The GC and MS operate at high temperatures. Do not touch any parts of GC and MS until...
This Information Applies To: Agilent GC Automatic Liquid Samplers
Issue
Unusual issues can occur to a syringe: needle bending during sample injections.
Resolution
Ensure optimal system performance by considering the following options:
Vials and caps
Agilent vials and caps should be used...
This Information Applies To: 7890 or 8890/8860 GCs with a split/splitless inlet, MMI, PTV inlet, or volatile interface inlet.
Issue
Check the split vent trap line for contamination or an abnormal restriction causing high inlet pressure.
Items Required
No-hole ferrule ( 5190-4054 ) and column...
This Information Applies To: Agilent GC Systems
Issue
The GC carrier gas actual pressure exceeds its pressure setpoint.
Resolution
Split/splitless inlet, MMI inlet, PTV inlet, or volatiles interface
Check the split vent trap line for contamination or an abnormal restriction. Run the split...
This Information Applies To: Agilent GC Systems
Issue
The Agilent Gas Clean Carrier Gas Filter delivers faster stabilization times for increased GC productivity, removes oxygen, moisture, and hydrocarbons from the carrier gas for MS applications, and provides ultimate GC column protection. The...