Hello
How can I set up method or sequence so pressure (or other parameters) can be saved and displayed with chromatogram?
I can find option to save it (like in classic Chemstation) and In DA I cannot see any instrument signals only detector signal.
Regards
Tomasz Kubowicz
Hello,
I added tags to this post to increase visibility.
To view the instrument traces in Data Analysis, you need to make sure that the injection tree properties are setup properly. To access the injection tree properties, left in the injection tree on the left hand side of Data Analysis to select it. Then in the top ribbon pane select the Injection Tree tab. Then left click on the icon outlined in the below screenshot:
In the Injection Tree Properties, there are two options specific to displaying instrument traces: 1) 'Show instrument trace selector' and 2) 'Show instrument traces in injection details':
Below in a screenshot showing what each of those options add to the injection tree:
Either option should allow you to display instrument traces in Data Analysis.
Hopefully this helps.