Hello
Is there any built-in chart in OL to generate residuals plot for calibration curve?
Or it needs to be created from scratch?
I've been trying to find it in reports items (charts/calibration etc.) but it looks like there isn't one.
Also in calibration curve tab even if I have few injections per level I can see only one residual value (Residual per level, rel residual, rel residual%) - I guess it is average. Is there any chance to display all of them (for every single injection)?
Regards
Tomasz Kubowicz
Hello,
I added tags to this post to increase visibility.
There is not built-in chard for displaying residuals for a calibration curve.
For your second question it depends on how you have your curve calculation set in your processing method. The options are 'From average per level' or 'From individual calibration points'.
If you have 'From average per level' selected then you will only see one Residual per level, rel. residual, and rel. residual % value displayed for each level. These values would be the based off of the average of the calibration points for that level.
If you have 'From individual calibration points' selected then you will see Residual per point, rel. residual, and rel. residual % values displayed for each calibration point.
Hopefully that helps.