    Openlab cds 2.3 calibration standards average and RSD table in reporting

    waynecheng

      Hi

      I've setup a linear calibration acquisition with 1 level and 5 repeats. I wanted to evaluate the RSD of the peak areas, RT and amount.

       

      I have an issue with reporting my calibration results. The calibration summary tab pulls my calibration curves 1 by 1, and does not provide a summary table for the calibration

       

      I came across this link, which has the report format that I need but the .rdl files cannot be read by v2.3

      https://community.agilent.com/docs/DOC-1380

       

      Any help is appreciated!

        • Re: Openlab cds 2.3 calibration standards average and RSD table in reporting
          ryoboyle

          Hello,

           

          I updated the tags on this post to increase visibility.

           

          Yes, those templates attached to the post you linked were designed for OpenLab ChemStation and OpenLab EZChrom software and are not compatible in OpenLab CDS 2.3.

           

          To give you an idea of how the summary table is set up, below are screenshots of the Peaks and Repeating and Filtering properties on the table. If you create a table on your report with the same properties then you should be able to reproduce the summary table you want.

           

           

           

          Hopefully this helps.