    Fault light flashing

    carolpollard

      Hi,

       

      The fault light on our injection tower keeps flashing. I counted it and its 4 flashes. Anyone knows what this means? Been trying to find it online but have had no luck!

       

      Thanks,
      Caroline

        • Re: Fault light flashing
          james_jenkins

          Hi carolpollard,

          The injector fault lights do have some meaning. I have pasted a chart below. Its fairly generic but does cover all models of GC ALS towers that I can think of. In your case this is a plunger error. This can be caused by a syringe that is binding. If you have changed the syringe and still get the error, then the tower needs service as there is a problem with the plunger motor circuitry. I hope this helps.

          Regards

          Jamestower.jpg

          1 person found this helpful