1 Reply Latest reply on Jan 28, 2019 1:30 PM by ryoboyle

    How to print the modified sequence file after the run completion in OpenLAB CDS 2.3

    ragu

      Hello,

      I need to print the sequence file after modification during run in OpenLAB CDS 2.3.

       

      I tried to load the resultset sequence file as sequence in AQ after run completion, and tried to print. its not allowed to print without saving. Please help.

        • Re: How to print the modified sequence file after the run completion in OpenLAB CDS 2.3
          ryoboyle

          Hello,

           

          You are correct that in currently you need to save sequence templates before printing them from Acquisition, so meaning that each sequence you import from a result set would need to be saved as a new acquisition sequence template before printing.

           

          To avoid this, what you would need to do is to build a sequence summary report template in Data Analysis that will output the same information as the Acquisition report. You should be able to do so by using the Sequence>Sample Summary and Sequence Audit Trail report items.

           

          Hopefully this helps.