2 Replies Latest reply on Dec 10, 2018 5:00 AM by mr.white

    INTELLIGENT REPORTING

    agarg

      Hi,

      I have added 5 named peaks in my calibration table but results are based on only 1 calibration curve.

       

      While I am reporting, results for all calibration peaks coming to the report rather than just I wanted for main peak. How I can sort a claibration table to include results (peak area, amount) for only 1 calibration peak.

      Thanks

        • Re: INTELLIGENT REPORTING
          wydnaa

          If you want to report only the results from the main peak, then you need to apply a filter to the compound results table to only show the content that you want.  So, for example, if you were measuring Paracetamol the filter would look something like this:

           

          Expression:     Compound_Name LIKE "*Paracetamol*"

          Operation:       =

          Value:              =True

           

          The expression is case-sensitive and uses wildcards to filter for anything that contains "Paracetamol"

           

          /Andy