    Intelligent Reports - a table with only one peak?

    michalab

      Hello,

      I am trying to create a table in one of my Intelligent Reports that only calls out one specific peak in the column 'Peak Name'. So far I can only find a way to have all of the peaks listed when I create column 'Peak Name', I'm trying to get only one specific peak to be listed (while keeping all of the other peaks still labeled in the chromatogram so I can still call them all out in another table in the report). How can I do this?

      Thank you,

      Michal

        • Re: Intelligent Reports - a table with only one peak?
          ryoboyle

          Hello,

           

          You can add a filter into the table's properties to only display results for compounds with a specific compound name. For example, below I have a filter set up in for a report table so that it only displays results for biphenyl:

           

          You can access the table properties by right-clicking on the table and selecting properties. You can also apply filters to groups.