    Openlab 2.3 calibration level custom calculation syntax

    camelia

      I am using 6 levels of a calibration standard that contains Compound A and Compound B. I want to extract the calibration amount of level 4 (the theoretical concentration entered in the calibration table) for Compound A and Compound B in order to calculate: sample amount Compound A / calibration level 4 amount Compound A *100. What is the syntax to use for this custom calculation please?

        • Re: Openlab 2.3 calibration level custom calculation syntax
          ryoboyle

          Hello,

           

          I added tags to this post to increase visibility.

           

          Do you want the amounts values entered into the calibration table to be displayed on a report? You could do this either by adding the DataAnalysis Information Report Item to the report or by adding a Single Data Analysis Method Table Row and specify the for which compound you want this displayed for:

           

           

          Unfortunately, with both of these methods you will not be able to call these values to be used in any custom calculations.

           

          Beyond that, you might be able to set up a custom calculation that back calculates this value from a calibration sample data file for that level.

            • Re: Openlab 2.3 calibration level custom calculation syntax
              camelia

              Hello,

              What I would like to know what  syntax to use to obtain the calibration amount of the level 4 standards for Compound A and Compound B so I can use these data in a custom calculation - I do not want to use a report method for this.

              Please could you provide further information on how it might be possible to set up a custom calculation that back calculates this value from a calibration sample data file for that level.

              Thanks in advance

                • Re: Openlab 2.3 calibration level custom calculation syntax
                  ir_nerd

                  Hello Camelia,

                   

                  In Custom Calculation you can use the following expression to extract from your calibration sequence the calibration amount of your level 4:

                   

                  CurrentSequence.AllPeaksAndGroupsByName("Compound A").where(function (c) c.injection.sample.sample_type = 1).where(function (c) c.injection.sample.Sample_CalibrationLevel = 4). Select(function(c) c.Compound_CalibrationAmount). first

                   

                  Your CC line must have a "Sequence" Scope, Visible and set as a double:

                   

                  Once you have linked your cc file in the processing method (or embed), reprocess all injections, and you should see the results in the injection results window under "summary":

                   

                  If so, the results can then be used in reporting taking the custom calculation results from the sequence scope:

                   

                   

                  Let me know if that works for you, I'll explain a little bit more the expression I gave you.