      Hi, I have a agilent GC 7890B with FID. currently  i am a facing a problem regarding this GC, problem is like that -

      all gases supply is ok, no leakage is found in detector portion. i have  cleaned the FID & Jet but flame is not burning.

      So kindly suggest me what to do ?

       

      Thanks

        • Re: 7890B GC flame ignition isue
          james_jenkins

          Hi mgupta*9615,

          Please let us know the detector parameters that you are using. In most cases the issue is with the Jet. Jets are hard to clean properly and the replacement should be the next step.

          Regards

          James

           

          Typical parameters

          Air 400 ml/min

          H2 40 Ml/min

          Makeup, with today's detectors you do not need much, 5 to 10 ml/min if needed at all.

          Temperature 300 C

           

          If it doesnt light with the above change the jet, check the column connection for leaks.

          • Re: 7890B GC flame ignition isue
            mgupta*9615

            Hello James,

            Thanks for your valuable response & support. i'll try it .

            • Re: 7890B GC flame ignition isue
              mr.white

              I had the same Problem.

              I just changed the N2 Makeup Gas from 35 to 25. 11 you will loose so u get 14 at the detector.

              Agilent recommend these settings too. Air to H2 supposed to be 1 to 10.

              Check on the top of your device if it glowing while it say " ignition".

              Also you can hear the igniting sound.

              You can also check the outcome of your column. Just activate the flow and check with a glass of water at the end of the column if you see small bubbles.

               

              Hope I can help