Re: 5973A MSD HPIB communication
m4estiMay 23, 2018 10:30 AM (in response to ronin)
Hi,
What model hpib card do you have in the PC? Make sure the bus address is 30 in io libs. You need to scroll down to get this.
Mitch
-
Re: 5973A MSD HPIB communication
ronin May 23, 2018 12:42 PM (in response to m4esti)82350 (82650-66512)
-
Re: 5973A MSD HPIB communication
ronin May 23, 2018 12:44 PM (in response to ronin)Where can I change the Bus address?
-
Re: 5973A MSD HPIB communication
m4estiMay 23, 2018 1:48 PM (in response to ronin)
1 person found this helpful -
Re: 5973A MSD HPIB communication
m4estiMay 23, 2018 1:49 PM (in response to m4esti)
What rev of IO Libs are you running?
Mitch
-
Re: 5973A MSD HPIB communication
ronin May 23, 2018 2:05 PM (in response to m4esti)Supplyd with G1701DA D.03.00
Also I tried Agilent IO Libraries Suite v 16.2
This mentioned that the hardware is connected on adress 20
-
Re: 5973A MSD HPIB communication
ronin May 23, 2018 2:09 PM (in response to ronin)If i changed the bus address to 30, gives the same message.
Is this the same address as mentitioned in the manuals of the 5973 set to 20?
-
Re: 5973A MSD HPIB communication
ronin May 23, 2018 2:14 PM (in response to ronin)In the edit menu it is mentioned the serialnumber of the MSD.
The edit menu is little bit different
Serialnumber
Visa Interfac Name: GPIB0
Logical Unit: 7
Bus Address: 30 20?
Sytemconrol ^
-
Re: 5973A MSD HPIB communication
ronin May 23, 2018 2:28 PM (in response to ronin)I changed again the name of GPIB0 to hp82341 and bus adress to 30
It's now working!
So far, Thanks a lot, you helped me to the right direction
Can you explain me about the address 30?
Dip switches are on 20
MSD Sw Configuration is on 20
-
-
-
-
-
-
-
-