Re: 6890N working with windows 7
valentinrusuFeb 20, 2018 1:10 PM (in response to jeremyan)
Hello, what will be key in answering your questions is the type of software you're using. If you can turn on your acquisition software, please navigate to Help > About (or Review Revision History) and post the software name and Revision.
-
Re: 6890N working with windows 7
james_jenkinsFeb 20, 2018 1:17 PM (in response to jeremyan)
Hi jeremyan,
For Windows 7 you need MSD Chemstation E02.01 or higher for software with the matching IO libraries from the install disk. I would check the basics first with the PC, are you able to ping the 6890N? If so, basic communication would be working. I would suggest following the installation guide I have attached. The concern here with the 5973N is the display. This could be a display only issue, or could be the smart card and display. The problem with no display is that you do not know aht IP address has been used with the MSD. If you are lucky, its the most common 10.1.1.102.
Regards
James
I have also updated the tags for better visibility .
-
G1701-90066.pdf 3.1 MB
-
Re: 6890N working with windows 7
jeremyan Feb 21, 2018 4:43 AM (in response to james_jenkins)I am not sure of the chemstation version number, it is the same version that came with the 6890N for the original XP machine. I can ping the GC but not the MSD. I was told that the IP for the MSD is the default (10.1.1.102) but I have no way of confirming that. If I need a newer version of the chemstation could you provide me the instructions for obtaining that? Also would you be able to provide me with troubleshooting instructions and repair part numbers to order so I can get the MSD operational?
-
Re: 6890N working with windows 7
james_jenkinsFeb 21, 2018 5:11 AM (in response to jeremyan)
The software needed would be a purchasable upgrade and there are a couple of options such as moving forward to the MassHunter platform. My thought on the MSD is that it needs a new LUI (Display) and smart card as they should always be replaced as a pair when there is a display issue. There are other issues that go beyond what we can easily cover in this public forum. I would suggest calling 1-800-227-9770 opt 3 for a live discussion with one of our Mass Spec Service Engineers.
Regards
James
-
Re: 6890N working with windows 7
valentinrusuFeb 22, 2018 7:05 AM (in response to jeremyan)
Hi jeremyan, if you're able to ping the GC, you might try configuring an instrument in MSD Configuration with the GC but no MSD.
As far as your actual revision of ChemStation, we would recommend you find out exactly what you have. If it's E.02.01 or newer, we might be able to help you through this forum. If you could at least open your Agilent MSD Configuration (it should be a shortcut on the desktop), and then click on Help > Review Revision History..., that should tell us the revision. The window you get will look like this:
In my case, I'm using E.02.02 SP2.
To obtain a newer version than what you have, you would contact your local sales representative. If you don't have that contact information, you can call in to our support and sales toll-free number in the U.S., 800-227-9770. You can also navigate to the Contact Us page on agilent.com: https://www.agilent.com/en-us/contact-us/page for more information.
-
-
-
Re: 6890N working with windows 7
caduceon Feb 22, 2018 7:40 AM (in response to jeremyan)I wouldn't worry too much about the software version right now. Every version of Chemstation, from A to E will still connect to the 5973. The only exception to that is that a 5973A will not easily connect with version E, you have to use Version D or earlier for those instruments. Is the mass spec hooked up via GPIB cables? or is it a usb to GPIB adapter? In order for the mass spec to connect, you'll have to double check the IP, then configure the gpib adapter. The GC's generally connect without issue, but the older mass specs are more finicky. There are a few specific parameters that you have to configure for the connection to work. In the IO libraries, for the SICL interface name, you have to name it "hp82341".
Now, the other thing is that if the screen is not coming on, then identifying the IP it is using will be more difficult. Although, that may indicate a hardware issue with the mass spec. Does the turbo pump/diffusion pump turn on? does it attempt to pump down? If it does not do those things than the mass spec will need servicing for sure.
-
Re: 6890N working with windows 7
valentinrusuFeb 22, 2018 7:45 AM (in response to caduceon)
The software revision is crucial to what will and will not work on Windows 7. If it is not E.02.01 or later, it was not tested on that operating system and is not guaranteed to work. Spending time getting the communication working will be for naught if ultimately ChemStation can't run properly due to incompatibility. Agilent strongly recommend against trying to use an older, untested revision of MSD ChemStation on an untested operating system. Certainly no D revision or earlier, but even the early E revisions will likely not function properly.
-
Re: 6890N working with windows 7
jeremyan Feb 22, 2018 8:04 AM (in response to valentinrusu)Ok, I will be heading to the labs shortly as I work in multiple locations. I will check all of these points and get back to you with my findings.
Jeremy [personal information removed]
-
Re: 6890N working with windows 7
caduceon Feb 22, 2018 8:05 AM (in response to valentinrusu)Oh i agree, the earlier versions will most certainly not function "properly". But they WILL install, and they WILL connect to the mass spec. Will they run after that? Very possibly not. But i figure first things first. On a related note, The old 5973As did have a hidden button you could press to manually reset the networking card, which would reset the IP address to the original default value. I do not know if the model that Jeremyan uses, still has that function or not. Possibly power cycling both the GC and the Mass Spec simultaneously can help also.
-
Re: 6890N working with windows 7
james_jenkinsFeb 22, 2018 10:14 AM (in response to caduceon)
Hi caduceon,
One minor point of clarification. The older 5973A did have the reset switch, but its only function was to reset the smart card which was similar to a power cycle. The 5973A was GPIB only and addressing was handled by dip switch settings. It looks like jeremyan is referring to a 5973N as he mentions the LUI display and that would be a full LAN/IP instrument.
-
Re: 6890N working with windows 7
jeremyan Feb 22, 2018 12:43 PM (in response to james_jenkins)That is correct sir, I am currently at the lab both the 6890 and the 5973 are of the N persuasion. As for the chem station version it is currently D.01... so it is definitely older than recommended. I do have another possible option available as there is another pc with windows seven OS that has masshunter B.05.... version installed and running a newer GC\MSD. would it be possible to run the old 6890N from this machine? I am assuming the correct IO library and a few other settings would need attention, but would it be feasible? Also just to cover the bases could you put me in touch with some one who could give me a quote for the software I would need to get this GC running on it's own computer?
Jeremy [personal information removed]
-
Re: 6890N working with windows 7
james_jenkinsFeb 22, 2018 11:52 AM (in response to jeremyan)
Hi Jeremy,
The D01.XX would definitely have some problems with functionality on Win 7. The MassHunter B05.XX version will not be an option as we only started supporting the 5973N on B07.01 and later versions. I will send your information to our sales dept for software information. If you are not able to ping the MSD using basic commands , there is a real possibility that the smart card has an issue.
Regards
James
-
Re: 6890N working with windows 7
james_jenkinsFeb 22, 2018 11:58 AM (in response to jeremyan)
Hi Jeremy,
I just left a voicemail on your cell , feel free to email me the requested information direct with the email I left in the message.
James
-
-
-
-
-