Re: Documentation for 7673A Controller Module
james_jenkinsJan 29, 2018 10:53 AM (in response to fomalhaut)
Hi fomalhaut,
I updated the tags on this post for better visibility . The combination you have will not work without the GPIB (HPIB) card in the 7673 controller box. The G1800A GCD only supported the use of GPIB. The Inet connection is there on the GC portion , but inet was meant for simple intergrator connections and not GCD control software or the GCD. As you can tell the GCD is a 5890 with a MSD attached on the side as one box. Very limited in options and features. I looked at the GCD product support plan and the 7673A was not a match to the GCD. The 7673B with GPIB was the supported workable configuration.
Its good to hear of one of these instruments still around as they introduction was in 1994. Unfortunately the instrument is well out of support with very little to no parts available.
Regards
James
-
Re: Documentation for 7673A Controller Module
fomalhaut Feb 1, 2018 12:51 PM (in response to fomalhaut)Yes thank you for your prompt response. this is the info I had been expecting however disappointing it is for my situation. Yes this instrument has been a workhorse around here and I too am amazed it's held up so well. Although I operate much newer instruments from Agilent, I like older equipment as well as it's much simpler to understand and maintain as long as we keep spares around.
I have an additional problem somewhat related to this one. The original HPIB board in use for this GCD is an ISA card. I have acquired a used PCI GPIB board by National Instruments which can be installed in this same old PC or an updated PC. As I fear the PC will itself die soon leaving me without a usable HPIB pc interface, I want to try to get this PCI card to take it's place. Do you think that is even possible? I haven't yet tried it but have NI software for its operation which should run under Win95.
If you think I should post this as a separate question first please let me know. I'm new to this process although I'm "old school".
-
Re: Documentation for 7673A Controller Module
james_jenkinsFeb 1, 2018 1:43 PM (in response to fomalhaut)
Perfectly ok to post in this thread. I have had customers try other GPIB boards and have seen a few from National Instruments. I do not recall any success stories with any cards other than the HP/Agilent cards. I have not personally configured one a third party one either, but perhaps someone else on the community has and can comment. While not supported, certainly to keep a reliable instrument working, I know exactly why you would want to explore doing this. We do have a PCI GPIB card, but I do not believe we have any drivers that support older operating systems. The earliest would be XP to my knowledge. Hopefully someone else can jump in here with information. Certainly, if you have success,please keep us up to date.
Regards
James
-
-
Re: Documentation for 7673A Controller Module
james_jenkinsFeb 9, 2018 5:06 AM (in response to fomalhaut)
Hi fomalhaut,
I went ahead and flagged this post as “Assumed Answered”. If you still need help though, please let us know and we would be happy to continue working with you. If the issue is resolved please feel free to update us or mark correct answer.
Regards
James