    mass 32 is not detected

    bperezf

      I am working with a GC/MS 6890 5973.

      And in the last tune there is not response for the mass 32. Could you imagine any reason?

       

      Thanks

        • Re: mass 32 is not detected
          james_jenkins

          I have updated the tags for better visibility. Mass 32 is oxygen  which is one of the things we work to minimize in a GCMSD. The less air in a system, the better the mean free path. Sounds like a tight system, but I would need more information on why you are concerned about this.  Usually lack of 32, 28 and 18 is a good thing.

          James