    Calibration curves ICPMS

    timharcourt

      Hi,

       

      I am currently carrying out validations for USP 233 for ICPMS.

       

      Regarding calibration curves I am using a 5 point calibration curve starting from Blank (Matrix). Should the calibration curve on MassHunter be set to Offset or Ignore?

       

      Many thanks,

       

      Tim

        • Re: Calibration curves ICPMS
          james_jenkins

          Hi Tim,

          I updated the tags on this for better visibility .

          James

          • Re: Calibration curves ICPMS
            gchaplain

            Hi Tim,

             

            As a rule of thumb I would always ignore the intercept.

             

            Reason being is that you will be including a result that you have not measured.

             

            You are using a blank for your lower end concentration, meaning that the blank result (if there is one) will be included in the calibration.

             

            NOTE: This is a statistically debated area. Personally, though the calibration curve you produce may not be straight line linear at the lower end concentrations (due to noise interfearance).

             

            Hope this helps.

             

            Grant