15 Replies Latest reply on Jun 11, 2018 6:59 AM by yaguo

    Software MassHunter Workstation Qualitative Ananlysis

    ulisalves

      Gostaria de um auxílio pois ao instalar e depois executar o software MassHunter Workstation Qualitative Ananlysis sempre abre a seguinte mensagem de erro:

       

      Agilent MassHunter Workstation Software Qualitative Analysis encountered an error. The program will attempt to continue ignoring this erro. However we recommend saving the data and restarting the program."

       

      Apesar de ter restartado o programa várias vezes, sempre ao abrir o software gera essa mensagem seguido de um botão "Ok". Após clicar, aparentemente o software funciona normalmente e gera um log, mas gostaria de saber o porque mostra essa mensagem de erro.

        • Re: Software MassHunter Workstation Qualitative Ananlysis
          dorth

          If this is a persistent error, it might be caused by a PC resource problem.  You can add RAM to the PC or try the same steps on another PC to see if it works.  You can also navigate to C:/Temp and look for a QualMemoryTrace.gz file, which contains diagnostic information about the error. 

          • Re: Software MassHunter Workstation Qualitative Ananlysis
            ulisalves

            Hi James

            This is the error message and Follows the log file. I'm a machine administrator.

             

             

            09/27/2017 11:11:57.006  ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~, ThreadID=1

            09/27/2017 11:11:57.162  Agilent MassHunter Workstation Software Qualitative Analysis, B.06.00, 6.0.633.0, 27/09/2017 11:11:57. START., ThreadID=1

            09/27/2017 11:11:57.162  ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~, ThreadID=1

            09/27/2017 11:12:02.631  Probe: C:\Program Files\Agilent\MassHunter\Workstation\Qual\B.06.00\help\pt-BR\QualDA.chm, ThreadID=1

            09/27/2017 11:12:02.631  Probe: C:\Program Files\Agilent\MassHunter\Workstation\Qual\B.06.00\help\pt\QualDA.chm, ThreadID=1

            09/27/2017 11:12:02.631  Probe: C:\Program Files\Agilent\MassHunter\Workstation\Qual\B.06.00\help\QualDA.chm, ThreadID=1

            09/27/2017 11:12:02.631  returning , ThreadID=1

            09/27/2017 11:12:11.912  , ThreadID=1

            09/27/2017 11:12:12.068  System.Reflection.TargetInvocationException: Exception has been thrown by the target of an invocation. ---> System.NotSupportedException: Specified method is not supported.

               at Agilent.MassSpectrometry.DataAnalysis.PSetAvailableReportTemplates.GetCultureSpecificTemplateFolderPath()

               at Agilent.MassSpectrometry.DataAnalysis.PSetAvailableReportTemplates..ctor()

               at Agilent.MassSpectrometry.DataAnalysis.PSetReportTemplateSelection..ctor()

               --- End of inner exception stack trace ---

               at System.RuntimeTypeHandle.CreateInstance(RuntimeType type, Boolean publicOnly, Boolean noCheck, Boolean& canBeCached, RuntimeMethodHandleInternal& ctor, Boolean& bNeedSecurityCheck)

               at System.RuntimeType.CreateInstanceSlow(Boolean publicOnly, Boolean skipCheckThis, Boolean fillCache, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(Type type, Boolean nonPublic)

               at System.RuntimeType.CreateInstanceImpl(BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(Type type, BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes)

               at System.Activator.CreateInstance(String assemblyString, String typeName, Boolean ignoreCase, BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes, Evidence securityInfo, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(String assemblyName, String typeName)

               at System.AppDomain.CreateInstance(String assemblyName, String typeName)

               at System.AppDomain.CreateInstanceAndUnwrap(String assemblyName, String typeName)

               at Agilent.MassSpectrometry.DataAnalysis.MethodSubGroup.get_Item(String key)

               at Agilent.MassSpectrometry.DataAnalysis.DAMethodBase.get_Item(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualDAMethod.get_Item(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.AppManager.GetMethodPSet(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.AppManager.GetMethodPset(String usageKey, Boolean raiseErrorIfInvalid)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.SetSettingsPSetReferences(Dictionary`2 keyToCurrent, Dictionary`2 keyToOriginal, DataItem settingsDataItem, ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.CreateSettingsNodes(SettingsEditorConfig config, ItemGroup itemGroupOfNodes, Dictionary`2 keyToCurrent, Dictionary`2 keyToOriginal, ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.ReadConfiguration(ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.Do(), ThreadID=1

            09/27/2017 11:12:12.068     at System.RuntimeTypeHandle.CreateInstance(RuntimeType type, Boolean publicOnly, Boolean noCheck, Boolean& canBeCached, RuntimeMethodHandleInternal& ctor, Boolean& bNeedSecurityCheck)

               at System.RuntimeType.CreateInstanceSlow(Boolean publicOnly, Boolean skipCheckThis, Boolean fillCache, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(Type type, Boolean nonPublic)

               at System.RuntimeType.CreateInstanceImpl(BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(Type type, BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes)

               at System.Activator.CreateInstance(String assemblyString, String typeName, Boolean ignoreCase, BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes, Evidence securityInfo, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(String assemblyName, String typeName)

               at System.AppDomain.CreateInstance(String assemblyName, String typeName)

               at System.AppDomain.CreateInstanceAndUnwrap(String assemblyName, String typeName)

               at Agilent.MassSpectrometry.DataAnalysis.MethodSubGroup.get_Item(String key)

               at Agilent.MassSpectrometry.DataAnalysis.DAMethodBase.get_Item(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualDAMethod.get_Item(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.AppManager.GetMethodPSet(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.AppManager.GetMethodPset(String usageKey, Boolean raiseErrorIfInvalid)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.SetSettingsPSetReferences(Dictionary`2 keyToCurrent, Dictionary`2 keyToOriginal, DataItem settingsDataItem, ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.CreateSettingsNodes(SettingsEditorConfig config, ItemGroup itemGroupOfNodes, Dictionary`2 keyToCurrent, Dictionary`2 keyToOriginal, ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.ReadConfiguration(ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.Do(), ThreadID=1

            09/27/2017 11:12:12.068  __________________________Application (applogic) state start___________________, ThreadID=1

            09/27/2017 11:12:12.068  Navigation mode 'NavigateByAssociation'. Tree ordered by 'AnalysisCentric'. Applocation Mode 'Default', ThreadID=1

            09/27/2017 11:12:12.068  Memory status : Enough memory available., ThreadID=1

            09/27/2017 11:12:12.068  Current Method   : C:\Program Files (x86)\MassHunter\Methods\B.06.00\Default.m, ThreadID=1

            09/27/2017 11:12:12.068  __________________________Application (applogic) log starts___________________, ThreadID=1

            09/27/2017 11:12:12.068  Writing last 8 application actions recorded, ThreadID=1

            09/27/2017 11:12:12.068  27/09/2017 11:12:04 : DA method changed to 'Default.m : C:\Program Files (x86)\MassHunter\Methods\B.06.00\Default.m' from 'QualDAMethod.m : ProgramDefaults.m' , ThreadID=1

            09/27/2017 11:12:12.068  27/09/2017 11:12:04 : (S)Execute - CmdInitializeApplication : Time : 00:00:00.5468855, ThreadID=1

            09/27/2017 11:12:12.068  27/09/2017 11:12:05 : (S)Execute - Scale Chromatograms Off : Time : 00:00:00.0156253, ThreadID=1

            09/27/2017 11:12:12.068  27/09/2017 11:12:05 : (S)Execute - Scale Spectra Off : Time : 00:00:00, ThreadID=1

            09/27/2017 11:12:12.068  27/09/2017 11:12:05 : (S)Execute - Scale Spectra Off : Time : 00:00:00, ThreadID=1

            09/27/2017 11:12:12.068  27/09/2017 11:12:05 : (S)Execute - Linked Navigation On : Time : 00:00:00, ThreadID=1

            09/27/2017 11:12:12.068  27/09/2017 11:12:05 : (S)Execute - Cycle Through Defined Colors : Time : 00:00:00, ThreadID=1

            09/27/2017 11:12:12.068  27/09/2017 11:12:07 : (S)Execute - MS Actuals Off : Time : 00:00:00.0156253, ThreadID=1

            09/27/2017 11:12:12.068  __________________________Application (Applogic) log ends_____________________, ThreadID=1

            09/27/2017 11:12:12.068  __________________________Application (applogic) state log ends_____________________, ThreadID=1

            09/27/2017 11:12:12.068 [ERROR] CommandModel.CommandHistory System.Reflection.TargetInvocationException: Exception has been thrown by the target of an invocation. ---> System.NotSupportedException: Specified method is not supported.

               at Agilent.MassSpectrometry.DataAnalysis.PSetAvailableReportTemplates.GetCultureSpecificTemplateFolderPath()

               at Agilent.MassSpectrometry.DataAnalysis.PSetAvailableReportTemplates..ctor()

               at Agilent.MassSpectrometry.DataAnalysis.PSetReportTemplateSelection..ctor()

               --- End of inner exception stack trace ---

               at System.RuntimeTypeHandle.CreateInstance(RuntimeType type, Boolean publicOnly, Boolean noCheck, Boolean& canBeCached, RuntimeMethodHandleInternal& ctor, Boolean& bNeedSecurityCheck)

               at System.RuntimeType.CreateInstanceSlow(Boolean publicOnly, Boolean skipCheckThis, Boolean fillCache, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(Type type, Boolean nonPublic)

               at System.RuntimeType.CreateInstanceImpl(BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(Type type, BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes)

               at System.Activator.CreateInstance(String assemblyString, String typeName, Boolean ignoreCase, BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes, Evidence securityInfo, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(String assemblyName, String typeName)

               at System.AppDomain.CreateInstance(String assemblyName, String typeName)

               at System.AppDomain.CreateInstanceAndUnwrap(String assemblyName, String typeName)

               at Agilent.MassSpectrometry.DataAnalysis.MethodSubGroup.get_Item(String key)

               at Agilent.MassSpectrometry.DataAnalysis.DAMethodBase.get_Item(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualDAMethod.get_Item(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.AppManager.GetMethodPSet(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.AppManager.GetMethodPset(String usageKey, Boolean raiseErrorIfInvalid)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.SetSettingsPSetReferences(Dictionary`2 keyToCurrent, Dictionary`2 keyToOriginal, DataItem settingsDataItem, ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.CreateSettingsNodes(SettingsEditorConfig config, ItemGroup itemGroupOfNodes, Dictionary`2 keyToCurrent, Dictionary`2 keyToOriginal, ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.ReadConfiguration(ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.Do()

               at Agilent.MassSpectrometry.CommandModel.CommandHistory.Invoke(ICommand cmd), ThreadID=1

            09/27/2017 11:12:12.084 Exception has been thrown by the target of an invocation. , ThreadID=1

            09/27/2017 11:12:12.662  Exception Location - App: AgtQual.exe, FileVersion: 6.0.633.0

            .....................Module: System.RuntimeTypeHandle

             

            Exception 1: System.Reflection.TargetInvocationException

            Message:  Exception has been thrown by the target of an invocation.

            Source:   mscorlib

            Site:     System.Object CreateInstance(System.RuntimeType, Boolean, Boolean, Boolean ByRef, System.RuntimeMethodHandleInternal ByRef, Boolean ByRef)

               *** Stack Trace ***

               at System.RuntimeTypeHandle.CreateInstance(RuntimeType type, Boolean publicOnly, Boolean noCheck, Boolean& canBeCached, RuntimeMethodHandleInternal& ctor, Boolean& bNeedSecurityCheck)

               at System.RuntimeType.CreateInstanceSlow(Boolean publicOnly, Boolean skipCheckThis, Boolean fillCache, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(Type type, Boolean nonPublic)

               at System.RuntimeType.CreateInstanceImpl(BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(Type type, BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes)

               at System.Activator.CreateInstance(String assemblyString, String typeName, Boolean ignoreCase, BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes, Evidence securityInfo, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(String assemblyName, String typeName)

               at System.AppDomain.CreateInstance(String assemblyName, String typeName)

               at System.AppDomain.CreateInstanceAndUnwrap(String assemblyName, String typeName)

               at Agilent.MassSpectrometry.DataAnalysis.MethodSubGroup.get_Item(String key)

               at Agilent.MassSpectrometry.DataAnalysis.DAMethodBase.get_Item(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualDAMethod.get_Item(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.AppManager.GetMethodPSet(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.AppManager.GetMethodPset(String usageKey, Boolean raiseErrorIfInvalid)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.SetSettingsPSetReferences(Dictionary`2 keyToCurrent, Dictionary`2 keyToOriginal, DataItem settingsDataItem, ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.CreateSettingsNodes(SettingsEditorConfig config, ItemGroup itemGroupOfNodes, Dictionary`2 keyToCurrent, Dictionary`2 keyToOriginal, ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.ReadConfiguration(ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.Do()

               at Agilent.MassSpectrometry.CommandModel.CommandHistory.Invoke(ICommand cmd)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.AppManager.Invoke(ICommand cmd)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualForm.SendSynchronouslyToAppLayer(QualCommandBase command, Boolean showOutstandingMsgBox)

             

            Exception 2: System.NotSupportedException

            Message:  Specified method is not supported.

            Source:   CoreTypes

            Site:     System.String GetCultureSpecificTemplateFolderPath()

               *** Stack Trace ***

               at Agilent.MassSpectrometry.DataAnalysis.PSetAvailableReportTemplates.GetCultureSpecificTemplateFolderPath()

               at Agilent.MassSpectrometry.DataAnalysis.PSetAvailableReportTemplates..ctor()

               at Agilent.MassSpectrometry.DataAnalysis.PSetReportTemplateSelection..ctor()

             

            *** Operating System Information ***

            Microsoft Windows NT 6.1.7601 Service Pack 1

             

            *** Loaded Modules ***

            Module 1: AgtQual.exe

              File version: 6.0.633.0

              Assembly version: 6.0.633.0

              File date: 2012-10-30 15:10:20

              Base address: 0x00000000012C0000

              Module memory size: 1654784 bytes

             

            Module 2: ntdll.dll

              File version: 6.1.7601.23889

              File date: 2017-08-11 03:36:37

              Base address: 0x0000000077890000

              Module memory size: 1744896 bytes

             

            Module 3: MSCOREE.DLL

              File version: 4.0.40305.0

              File date: 2010-11-21 01:23:48

              Base address: 0x000007FEF70C0000

              Module memory size: 454656 bytes

             

            Module 4: SYSFER.DLL

              File version: 14.0.2395.200

              File date: 2017-09-21 07:59:35

              Base address: 0x0000000075270000

              Module memory size: 581632 bytes

             

            Module 5: KERNEL32.dll

              File version: 6.1.7601.23889

              File date: 2017-08-11 03:34:55

              Base address: 0x0000000077670000

              Module memory size: 1175552 bytes

             

            Module 6: KERNELBASE.dll

              File version: 6.1.7601.23889

              File date: 2017-08-11 03:34:55

              Base address: 0x000007FEFD8D0000

              Module memory size: 434176 bytes

             

            Module 7: ADVAPI32.dll

              File version: 6.1.7601.23889

              File date: 2017-08-11 03:34:50

              Base address: 0x000007FEFF9F0000

              Module memory size: 897024 bytes

             

            Module 8: msvcrt.dll

              File version: 7.0.7601.17744

              File date: 2011-12-16 06:46:06

              Base address: 0x000007FEFE6D0000

              Module memory size: 651264 bytes

             

            Module 9: sechost.dll

              File version: 6.1.7600.16385

              File date: 2009-07-13 22:41:53

              Base address: 0x000007FEFF0E0000

              Module memory size: 126976 bytes

             

            Module 10: RPCRT4.dll

              File version: 6.1.7601.23889

              File date: 2017-08-11 03:35:02

              Base address: 0x000007FEFEFB0000

              Module memory size: 1232896 bytes

             

            Module 11: mscoreei.dll

              File version: 4.7.2114.0

              File date: 2017-08-29 22:56:16

              Base address: 0x000007FEF7020000

              Module memory size: 643072 bytes

             

            Module 12: SHLWAPI.dll

              File version: 6.1.7601.17514

              File date: 2010-11-21 01:24:22

              Base address: 0x000007FEFFAD0000

              Module memory size: 462848 bytes

             

            Module 13: GDI32.dll

              File version: 6.1.7601.23807

              File date: 2017-05-12 15:22:32

              Base address: 0x000007FEFF100000

              Module memory size: 421888 bytes

             

            Module 14: USER32.dll

              File version: 6.1.7601.23594

              File date: 2016-11-10 14:32:41

              Base address: 0x0000000077790000

              Module memory size: 1024000 bytes

             

            Module 15: LPK.dll

              File version: 6.1.7601.23807

              File date: 2017-05-12 15:22:33

              Base address: 0x000007FEFF320000

              Module memory size: 57344 bytes

             

            Module 16: USP10.dll

              File version: 1.626.7601.23894

              File date: 2017-08-16 12:29:31

              Base address: 0x000007FEFEA10000

              Module memory size: 831488 bytes

             

            Module 17: IMM32.DLL

              File version: 6.1.7600.16385

              File date: 2009-07-13 22:41:09

              Base address: 0x000007FEFE770000

              Module memory size: 188416 bytes

             

            Module 18: MSCTF.dll

              File version: 6.1.7601.23572

              File date: 2016-10-11 12:31:58

              Base address: 0x000007FEFF210000

              Module memory size: 1085440 bytes

             

            Module 19: VERSION.dll

              File version: 6.1.7600.16385

              File date: 2009-07-13 22:41:56

              Base address: 0x000007FEFC650000

              Module memory size: 49152 bytes

             

            Module 20: clr.dll

              File version: 4.7.2114.0

              File date: 2017-08-29 22:56:16

              Base address: 0x000007FEF4B40000

              Module memory size: 10350592 bytes

             

            Module 21: MSVCR120_CLR0400.dll

              File version: 12.0.52519.0

              File date: 2017-08-30 13:23:08

              Base address: 0x000007FEF79E0000

              Module memory size: 1011712 bytes

             

            Module 22: mscorlib.ni.dll

              File version: 4.7.2114.0

              File date: 2017-09-26 19:49:06

              Base address: 0x000007FEECA30000

              Module memory size: 22536192 bytes

             

            Module 23: ole32.dll

              File version: 6.1.7601.23889

              File date: 2017-08-11 03:35:01

              Base address: 0x000007FEFEAE0000

              Module memory size: 2080768 bytes

             

            Module 24: CRYPTBASE.dll

              File version: 6.1.7601.23889

              File date: 2017-08-11 03:34:52

              Base address: 0x000007FEFD420000

              Module memory size: 61440 bytes

             

            Module 25: uxtheme.dll

              File version: 6.1.7600.16385

              File date: 2009-07-13 22:41:56

              Base address: 0x000007FEFBF50000

              Module memory size: 352256 bytes

             

            Module 26: System.ni.dll

              File version: 4.7.2114.0

              File date: 2017-09-26 19:49:15

              Base address: 0x000007FEEE900000

              Module memory size: 12722176 bytes

             

            Module 27: CRYPTSP.dll

              File version: 6.1.7601.23471

              File date: 2016-06-14 14:16:25

              Base address: 0x000007FEFCD90000

              Module memory size: 98304 bytes

             

            Module 28: rsaenh.dll

              File version: 6.1.7600.16385

              File date: 2009-07-13 22:43:15

              Base address: 0x000007FEFCA90000

              Module memory size: 290816 bytes

             

            Module 29: System.Core.ni.dll

              File version: 4.7.2114.0

              File date: 2017-09-26 19:56:02

              Base address: 0x000007FEEC070000

              Module memory size: 10219520 bytes

             

            Module 30: AgtQual.ni.exe

              File version: 6.0.633.0

              File date: 2017-09-27 11:11:39

              Base address: 0x000007FEF13D0000

              Module memory size: 3317760 bytes

             

            Module 31: System.Drawing.ni.dll

              File version: 4.7.2114.0

              File date: 2017-09-26 19:57:34

              Base address: 0x000007FEF0650000

              Module memory size: 2007040 bytes

             

            Module 32: System.Windows.Forms.ni.dll

              File version: 4.7.2114.0

              File date: 2017-09-26 19:57:45

              Base address: 0x000007FEEB140000

              Module memory size: 15900672 bytes

             

            Module 33: System.Configuration.ni.dll

              File version: 4.6.1055.0

              File date: 2017-09-26 19:56:05

              Base address: 0x000007FEF6AB0000

              Module memory size: 1179648 bytes

             

            Module 34: System.Xml.ni.dll

              File version: 4.7.2114.0

              File date: 2017-09-26 19:55:51

              Base address: 0x000007FEEE060000

              Module memory size: 9007104 bytes

             

            Module 35: nlssorting.dll

              File version: 4.6.1055.0

              File date: 2015-11-05 21:34:06

              Base address: 0x000007FEFA310000

              Module memory size: 90112 bytes

             

            Module 36: RpcRtRemote.dll

              File version: 6.1.7601.17514

              File date: 2010-11-21 01:24:01

              Base address: 0x000007FEFD4D0000

              Module memory size: 81920 bytes

             

            Module 37: api-ms-win-core-xstate-l2-1-0.dll

              File version: 10.0.10586.788

              File date: 2017-01-18 13:36:12

              Base address: 0x000007FEFA750000

              Module memory size: 12288 bytes

             

            Module 38: clrjit.dll

              File version: 4.7.2114.0

              File date: 2017-08-29 22:56:16

              Base address: 0x000007FEF12B0000

              Module memory size: 1130496 bytes

             

            Module 39: shell32.dll

              File version: 6.1.7601.23893

              File date: 2017-08-15 12:29:44

              Base address: 0x000007FEFD940000

              Module memory size: 14196736 bytes

             

            Module 40: comctl32.dll

              File version: 5.82.7601.18837

              File date: 2015-04-24 15:17:26

              Base address: 0x000007FEF9EB0000

              Module memory size: 655360 bytes

             

            Module 41: comctl32.dll

              File version: 6.10.7601.18837

              File date: 2015-04-24 15:12:05

              Base address: 0x000007FEFBB00000

              Module memory size: 2048000 bytes

             

            Module 42: gdiplus.dll

              File version: 6.1.7601.23894

              File date: 2017-08-16 12:29:12

              Base address: 0x000007FEFB8E0000

              Module memory size: 2191360 bytes

             

            Module 43: WindowsCodecs.dll

              File version: 6.2.9200.21830

              File date: 2016-04-09 00:52:25

              Base address: 0x000007FEFB5D0000

              Module memory size: 1445888 bytes

             

            Module 44: dwmapi.dll

              File version: 6.1.7601.18917

              File date: 2015-07-09 14:58:26

              Base address: 0x000007FEFB780000

              Module memory size: 98304 bytes

             

            Module 45: System.Design.ni.dll

              File version: 4.6.1055.0

              File date: 2017-09-26 19:58:40

              Base address: 0x000007FEEA540000

              Module memory size: 12550144 bytes

             

            Module 46: oleaut32.DLL

              File version: 6.1.7601.23775

              File date: 2017-04-17 12:37:29

              Base address: 0x000007FEFE930000

              Module memory size: 892928 bytes

             

            Module 47: System.Data.ni.dll

              File version: 4.7.2114.0

              File date: 2017-09-26 19:58:31

              Base address: 0x000007FEE9C40000

              Module memory size: 9424896 bytes

             

            Module 48: System.Data.dll

              File version: 4.7.2114.0

              Assembly version: 4.0.0.0

              File date: 2017-08-29 22:56:16

              Base address: 0x000007FEE98E0000

              Module memory size: 3493888 bytes

             

            Module 49: WS2_32.dll

              File version: 6.1.7601.23451

              File date: 2016-05-11 14:02:50

              Base address: 0x000007FEFFB50000

              Module memory size: 315392 bytes

             

            Module 50: NSI.dll

              File version: 6.1.7601.23889

              File date: 2017-08-11 03:35:01

              Base address: 0x000007FEFF630000

              Module memory size: 32768 bytes

             

            Module 51: CRYPT32.dll

              File version: 6.1.7601.23769

              File date: 2017-04-12 12:32:10

              Base address: 0x000007FEFD5B0000

              Module memory size: 1495040 bytes

             

            Module 52: MSASN1.dll

              File version: 6.1.7601.17514

              File date: 2010-11-21 01:24:22

              Base address: 0x000007FEFD590000

              Module memory size: 61440 bytes

             

            Module 53: System.Deployment.ni.dll

              File version: 4.7.2114.0

              File date: 2017-09-26 19:57:33

              Base address: 0x000007FEEF680000

              Module memory size: 2240512 bytes

             

            Module 54: profapi.dll

              File version: 6.1.7600.16385

              File date: 2009-07-13 22:41:53

              Base address: 0x000007FEFD580000

              Module memory size: 61440 bytes

             

            Module 55: System.Runtime.Serialization.Formatters.Soap.ni.dll

              File version: 4.6.1055.0

              File date: 2017-09-26 19:57:46

              Base address: 0x000007FEFACB0000

              Module memory size: 352256 bytes

             

            Module 56: Accessibility.ni.dll

              File version: 4.6.1055.0

              File date: 2017-09-26 19:55:38

              Base address: 0x000007FEFAE40000

              Module memory size: 77824 bytes

             

            Module 57: System.Numerics.ni.dll

              File version: 4.6.1055.0

              File date: 2017-09-26 19:56:05

              Base address: 0x000007FEFA500000

              Module memory size: 323584 bytes

             

            Module 58: diasymreader.dll

              File version: 14.0.1055.0

              File date: 2015-11-05 21:34:06

              Base address: 0x000007FEF0E10000

              Module memory size: 1142784 bytes

             

            Module 59: RichEd20.DLL

              File version: 5.31.23.1230

              File date: 2010-11-21 01:24:15

              Base address: 0x000007FEF8670000

              Module memory size: 647168 bytes

             

            Module 60: psapi.dll

              File version: 6.1.7600.16385

              File date: 2009-07-13 22:41:53

              Base address: 0x0000000077A60000

              Module memory size: 28672 bytes

             

             

            *** Threads ***

            Thread 1: Id: 5992

              Thread state:Running

              Base priority: 8, Current Priority: 10, Priority boost enabled: True

              Total processor time: 00:00:05.0937500, User processor time: 00:00:03.4531250, Priviledged processor time: 00:00:01.6406250

              Start address: 0x2005640512

            Thread 2: Id: 4940

              Thread state:Wait, Wait reason: UserRequest

              Base priority: 8, Current Priority: 8, Priority boost enabled: True

              Total processor time: 00:00:00, User processor time: 00:00:00, Priviledged processor time: 00:00:00

              Start address: 0x2005640512

            Thread 3: Id: 1872

              Thread state:Wait, Wait reason: UserRequest

              Base priority: 10, Current Priority: 10, Priority boost enabled: True

              Total processor time: 00:00:00, User processor time: 00:00:00, Priviledged processor time: 00:00:00

              Start address: 0x2005640512

            Thread 4: Id: 6032

              Thread state:Wait, Wait reason: UserRequest

              Base priority: 8, Current Priority: 10, Priority boost enabled: True

              Total processor time: 00:00:00, User processor time: 00:00:00, Priviledged processor time: 00:00:00

              Start address: 0x2005640512

            Thread 5: Id: 4960

              Thread state:Wait, Wait reason: EventPairLow

              Base priority: 8, Current Priority: 8, Priority boost enabled: True

              Total processor time: 00:00:00, User processor time: 00:00:00, Priviledged processor time: 00:00:00

              Start address: 0x2005640512

            Thread 6: Id: 1588

              Thread state:Wait, Wait reason: EventPairLow

              Base priority: 8, Current Priority: 8, Priority boost enabled: True

              Total processor time: 00:00:00, User processor time: 00:00:00, Priviledged processor time: 00:00:00

              Start address: 0x2005640512

            Thread 7: Id: 2500

              Thread state:Wait, Wait reason: ExecutionDelay

              Base priority: 8, Current Priority: 8, Priority boost enabled: True

              Total processor time: 00:00:00, User processor time: 00:00:00, Priviledged processor time: 00:00:00

              Start address: 0x2005640512

            Thread 8: Id: 5420

              Thread state:Wait, Wait reason: UserRequest

              Base priority: 8, Current Priority: 8, Priority boost enabled: True

              Total processor time: 00:00:00, User processor time: 00:00:00, Priviledged processor time: 00:00:00

              Start address: 0x2005640512

            Thread 9: Id: 2836

              Thread state:Wait, Wait reason: UserRequest

              Base priority: 8, Current Priority: 8, Priority boost enabled: True

              Total processor time: 00:00:00, User processor time: 00:00:00, Priviledged processor time: 00:00:00

              Start address: 0x2005640512

             

            , ThreadID=1

            09/27/2017 11:12:12.662  Date and Time: 27/09/2017  11:12

            Program: Agilent MassHunter Workstation Software Qualitative Analysis    B.06.00

            Build Configuration: Release

            Message: Exception has been thrown by the target of an invocation.

            Stack Trace:    at System.RuntimeTypeHandle.CreateInstance(RuntimeType type, Boolean publicOnly, Boolean noCheck, Boolean& canBeCached, RuntimeMethodHandleInternal& ctor, Boolean& bNeedSecurityCheck)

               at System.RuntimeType.CreateInstanceSlow(Boolean publicOnly, Boolean skipCheckThis, Boolean fillCache, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(Type type, Boolean nonPublic)

               at System.RuntimeType.CreateInstanceImpl(BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(Type type, BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes)

               at System.Activator.CreateInstance(String assemblyString, String typeName, Boolean ignoreCase, BindingFlags bindingAttr, Binder binder, Object[] args, CultureInfo culture, Object[] activationAttributes, Evidence securityInfo, StackCrawlMark& stackMark)

               at System.Activator.CreateInstance(String assemblyName, String typeName)

               at System.AppDomain.CreateInstance(String assemblyName, String typeName)

               at System.AppDomain.CreateInstanceAndUnwrap(String assemblyName, String typeName)

               at Agilent.MassSpectrometry.DataAnalysis.MethodSubGroup.get_Item(String key)

               at Agilent.MassSpectrometry.DataAnalysis.DAMethodBase.get_Item(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualDAMethod.get_Item(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.AppManager.GetMethodPSet(String usageKey)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.AppManager.GetMethodPset(String usageKey, Boolean raiseErrorIfInvalid)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.SetSettingsPSetReferences(Dictionary`2 keyToCurrent, Dictionary`2 keyToOriginal, DataItem settingsDataItem, ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.CreateSettingsNodes(SettingsEditorConfig config, ItemGroup itemGroupOfNodes, Dictionary`2 keyToCurrent, Dictionary`2 keyToOriginal, ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.ReadConfiguration(ActionItemCollection actions)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualCommandBase.Do()

               at Agilent.MassSpectrometry.CommandModel.CommandHistory.Invoke(ICommand cmd)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.AppManager.Invoke(ICommand cmd)

               at Agilent.MassSpectrometry.DataAnalysis.Qualitative.QualForm.SendSynchronouslyToAppLayer(QualCommandBase command, Boolean showOutstandingMsgBox)

            Message: Specified method is not supported.

            Stack Trace:    at Agilent.MassSpectrometry.DataAnalysis.PSetAvailableReportTemplates.GetCultureSpecificTemplateFolderPath()

               at Agilent.MassSpectrometry.DataAnalysis.PSetAvailableReportTemplates..ctor()

               at Agilent.MassSpectrometry.DataAnalysis.PSetReportTemplateSelection..ctor(), ThreadID=1

            • Re: Software MassHunter Workstation Qualitative Ananlysis
              james_jenkins

              Hi ulisalves,

              I have attached a file that may help.The document references Quant B05.00 but is relevant to Qual and Quant of later versions. Please take a look and see if it helps.

              James

              • Re: Software MassHunter Workstation Qualitative Ananlysis
                m4esti

                Ok, when you load Qual, do you right click and 'Run as Administrator"?