AnsweredAssumed Answered

6850 Oven Temperature Ramp Issues

Question asked by aline.castilla-vitkovsky on Aug 7, 2017
Latest reply on Oct 6, 2017 by james_jenkins

We are having issues with our 6850, if a ramp (°C/min) exceeds 30°C/min the oven temperature lags being starting at 108°C, by the time the oven temperature set point reaches ~200°C the actual final oven temperature differs by ~30°C.    

The retention time of the compounds did not change as they all come out before the ramp starts. Based on the consistency of the runs I think that if we had a compound with a RT within the ramp it might not change. Further testing is needed to confirm.

I ran one sample six times for each method, here are the results:

 

Run #

Method (80°C for 5.6 min then ramp to 200°C)

Not Ready Displayed at

Oven Temp Actual at end of Run

Oven Temp Set Point at end of Run

1

Ramp @ 50°C/min for a total run time of 8 min

110°C

166°C

197°C

2

108°C

165°C

197°C

3

108°C

165°C

197°C

4

108°C

165°C

197°C

5

108°C

165°C

197°C

6

108°C

165°C

197°C

1

Ramp @ 25°C/min for a total run time of 10.5 min

NA

197°C

198°C

2

NA

198°C

199°C

3

NA

197°C

198°C

4

NA

197°C

198°C

5

NA

197°C

198°C

6

NA

197°C

198°C

 

Any clue as to what the problem could be?