    Nitrogen consumption for 7200B QTOF?

    jonsun

      The specifications states that the system uses 2-3 L/min of N2 for purging/cooling of electronics. Which electronics does this cover, and is it a constant flow, i.e. also when the instrument is idle? Or is it for cooling of the RIS chamber when switching ion source? 

        • Re: Nitrogen consumption for 7200B QTOF?
          nickbharden

          Hi Jonsun,

          The N2 flow of 2-3L/min is to keep the high voltage TOF electronics clean, dry and at a steady temperature. These electronics are on all the time and therefore the N2 flow is needed even when the instrument is idle. Keeping the electronics on keeps the mass accuracy and resolution stable. The flow is constant.

          To purge the RIS we use up to 30L/min but this it only used when we are removing or inserting an ion source.

          Best regards,

          Nick

          • Re: Nitrogen consumption for 7200B QTOF?
            james_jenkins

            I went ahead and flagged the “Correct Answer” on this post.  If you still need help though, please let us know and we would be happy to continue working with you.