We noticed that the Setpoint contact of several (so far 8 out of 14) IPCMini opens for some minutes every 32 days, although the current is well below the Setpoint. Is this a software bug?

We noticed that the Setpoint contact of several IPCMini opens for some minutes every 32 days, although the current is well below the Setpoint.

The Setpoint is 1e-5A (or 10 uA).

After waiting for more than 30 minutes with that strange behaviour, the 5 kV high voltage was switched OFF and ON again and the Setpoint contact closed as was expected.

See the video below.

W

Is this a software bug?

Parents Reply Children
  • Thank you. While we're at it, there is another issue.

    The scaling of the IPCMini analog output can be set from menu.

    If the 230V power is cut and restored, the IPCMini will remember these 2 set values of the output scaling. Good.

    However, if somebody checks these values anytime after boot and presses "confirm", then these 2 values are set to zero.

    Thus, a second checking of the values reveals this.

    Here is a more detailed information as "picture story" combined with measurements as PDF from Klystron #8 from Thales.

     

    Best regards,

    Dr. Alexander Schnase