i have a VSMR15 and the spec tube light has been on for about a month and it is the color blue, any idea is to what is going on with it?
i have a VSMR15 and the spec tube light has been on for about a month and it is the color blue, any idea is to what is going on with it?
Hello
Here is an answer I received directly from the vacuum division:
The “spec tube” error is because the system has not achieved a proper
vacuum. It could be because the roughing pump is not running, or the test port
thermocouple is bad.
In the U.S., you can contact the Vacuum Division directly 800 882-7426, option 3
Regards,
Mitch