I want to use a Varian 979 as a pumping station to pump-clean a cold head. The cold head was filled with pure He4, so the leak rate would overload. Can I turn off the filament of the mass spectrometer to prevent overloading?
I want to use a Varian 979 as a pumping station to pump-clean a cold head. The cold head was filled with pure He4, so the leak rate would overload. Can I turn off the filament of the mass spectrometer to prevent overloading?
Hi isend,
We do not have any resources for Varian 979 leak detector. I will see if we have a resource for this question
Hello,
This is the answer I received from the vacuum division:
It won’t hurt the machine to use it as a pump, but why not just use the
roughing pump alone without the leak detector? It would pump much faster
directly attached.