Power Outage, now Openlab Aq and DA won't open but Control Panel will

Hey all our power went out this morning and since, we haven't been able to open the acquisition software, it just hang on the splash screen forever.  When we try Data analysis the splash screen opens then closes and software never opens.  In control panel it says HPLC Not Connected, but then if I try to Close Connection at the top it gives and error the HPLC is not active.

We were in the middle of a sequence when it happened.  

1260 Infinity II with OpenLab 2.4 (which apparently isn't supported any more?)

I called Agilent and they had me run the Software verification tool which passed everything.  

We've rebooted both computer and HPLC.  It seems to think the HPLC is still connected.

I am able to turn on the modules from Control Panel, but not open the other software.  

Any ideas?? Emergency situation 

Parents Reply Children
  • Hello,

    Look for the automation.log in C:\ProgramData\Agilent\LogFiles. You can open that log in notepad and look for any errors related to a firebird database or the secureworkarea. If you see any errors like that the secure work area database is corrupt. You would need to open a support ticket with your local Agilent support group, and they can help you recover the db or create a new one. 

    Marty

Was this helpful?