The calculation for limit of quantitation in Openlab field appears as value is "=First(Compound_LimitOfQuantitation)". I've added this field to a report template but there is no value output. Anyone found a way to get thisThanks
Find maintenance, best practice, and troubleshooting articles.
The calculation for limit of quantitation in Openlab field appears as value is "=First(Compound_LimitOfQuantitation)". I've added this field to a report template but there is no value output. Anyone found a way to get thisThanks
Hello,
The older help and learning had incorrect information on the fields LOD and LOQ, as they are not calculated in CDS. Those fields are for older EZChrom data reported in CDS 2.x. The latest help and learning updates have corrected this information. If you need to calculate LOD and LOQ, you can find a custom calculation file (Demo_LOD_LOQ.ccf) for LOQ and LOD on your media in the UCL folder. Also in later version the LOD and LOQ calculations are available in the IR
M8490-10009.zip\Setup\Tools\Support\UCL\CustomCalculatorExamples
Many thanks Martin, I have found the custom calculation so will try that out. The report items doesn't show calibration curves on this 2.4 version.
Yes, I can confirm that adding custom formula to the processing method and then reprcessing calculates both LOD / LOQ. thanks very much