Issue for GPC Addon with CDS 2.7

I have encountered two systems recently with the GPCSEC Addon for CDS 2.7.  In both cases, the GPC software is active in Data Analysis (I can create a GPC/SEC method and all the method setup sections are present).  However, in the GPC section, the peaks from narrow standards cannot be added (no right-click option), and none of the settings can be changes (Signal Choice, Type of Calibration, etc.).  After processing, a "GPC processing not permitted" error appears by the sample list.

I suspect the problem is in the license file but I can't confirm this.  When I view the licenses on working vs. non-working systems, I see slight differences in the GPC section of the license (AddonGPCSEC vs. GPCSECAddon).  Doesn't seem like a big difference but I wonder if it is related. 

Any help/suggestions are appreciated.

Thanks.

Parents Reply
  • Marty:

    Both installs were by Agilent FSE's.  I have details from one at this point about one fix. 

    The original install of 1.7 was configured but the FSE noted that some features were grayed out.  The installed version (1.7) was replaced with 1.6.1.  They also added the localhost login account to admin privileges in Windows 11, and then ran the configuration program. 

    I will have to check on these details because my message above indicated that 1.6.1 was installed.  So I don't know if the Windows settings were the key or not.  I can provide the contact information if you need more information.

    Thanks for responding quickly and helping with the process.