Data processing and reporting OpenLab CDS 01.17.27 GC

Hi All, I'm having trouble on OpenLab CDS 01.17.27.  I have analysed a sequence and want to display the data in a report (Compound Summary RT amount from Matrices - displaying peak area for named compounds). However, on previewing report I get 'no data available'. If I then go back to data analysis and click on individual injections and click 'Identify peaks, calculate + preview results' the info appears in preview - how do I apply this to all injections in the sequence at one go?  I think I'm missing something?

Parents
  • You will want to make sure you have both your data and the report template loaded in the Report section of the software. If you dock the left hand section of the software there should be a data and report template sub tab that you can load in both those items. To load multiple reports in one sequence you will want to single click on the first sample you want to load the report for, then hold Shift and double click the last sample you want to load a report for. The software will then load a report for those samples plus all of them in between that can then be exported in a single file.

Reply
  • You will want to make sure you have both your data and the report template loaded in the Report section of the software. If you dock the left hand section of the software there should be a data and report template sub tab that you can load in both those items. To load multiple reports in one sequence you will want to single click on the first sample you want to load the report for, then hold Shift and double click the last sample you want to load a report for. The software will then load a report for those samples plus all of them in between that can then be exported in a single file.

Children
  • Thanks L_Peach_49,  I will give it a try.

  • Hello,

    In your post you mentioned 'Identify peaks, calculate + preview results' as one of the steps. This button only exists in OpenLab Chemstation so I think that may be what you have installed. The OpenLab CDS 01.17.27 version number is not a valid version for that software. OpenLab CDS would be CDS 2.x where x is 0-7 and Chemstation would be C.01.xx where xx is 01-11. In Chemstation you should use the review tab to create reports from existing templates. The previous reply applies in the same manner for choosing the injections and template to be used. Please note the results must be processed in Chemstation before IR reports can be done. If you changed the data processing method make sure you reprocessed your results. Also note many of the canned reports only work with named peaks. If you are trying to report unknown peaks make sure to peak a report template that shows unknowns. 

    Marty Adams

  • Thanks Martin.  You are correct it is chemstation -should be C.01.07. I will try your suggestions.

Was this helpful?