Hi
How do I remove FID signal from the report for a dual TCD & FID GC, we use just TCD detector to analyses gas samples and this issue appear yesterday.
open lab rev. C01.05
Find maintenance, best practice, and troubleshooting articles.
Hi
How do I remove FID signal from the report for a dual TCD & FID GC, we use just TCD detector to analyses gas samples and this issue appear yesterday.
open lab rev. C01.05
Hello,
You could do it by setting up filters in the IR report to only display only one signal channel but it can also be done in the method as well. If you go under signal details in the method and using the drop down menu add only the signal channel you are interested to the table below. If the table is left empty then all channels will be reported. If you are reprocessing data and only want to see the selected channels make sure load using signal details is selected in the preferences. You should also check your acquisition method as you should be able to stop collecting the FID signal channel for new data.
Marty Adams
Hello,
You could do it by setting up filters in the IR report to only display only one signal channel but it can also be done in the method as well. If you go under signal details in the method and using the drop down menu add only the signal channel you are interested to the table below. If the table is left empty then all channels will be reported. If you are reprocessing data and only want to see the selected channels make sure load using signal details is selected in the preferences. You should also check your acquisition method as you should be able to stop collecting the FID signal channel for new data.
Marty Adams
Hi Adams,
Many thanks
After we checked the load using signal details in the preferences, We noticed we didn't selected this box ,
And the problem has been solved.
Dheya