OpenLab CDS reports: showing chromatography for analytes both when present or not present

Hello,

I'm trying to report the concentration of analyte using SIM mode. I have built a report that shows a table with the analyte and the ISTD, as well as chromatograms for each.
However, it only shows the analyte in the table (and the resulting chromatogram) when the analyte is present. I would like it to still display the analyte when not present: with zero concentration (and the resulting flat line in the chromatogram).
Does anyone have any ideas?

Using Agilent OpenLab Data Analysis Version 2.5 - Build 2.205.2.34

Thank you,
Rob

Parents Reply Children
  • Hi Andy,

    Thank you for the response. I found the selection to include missing compounds, but it only seems to show up in the table but not chromatogram. I am attaching example screenshots of what my problem is:

    In Report A, the analyte (1-bromo-2-propanol) is not present, but the ISTD is. In Report B both the analyte and ISTD are present.
    You can see that in Report A, only one chromatogram shows up; whereas in Report B both chromatograms are present.

    My goal is to basically that when 1-bromo-2-propanol is not present, to have a chromatogram with a flat line, showing no peak.

    Thank you,

    Robert

  • HI Robert,

    You're right, only the compound tables have an option to include missing compounds.  

    I'm now wondering what report template item you have selected for the display of the chromatograms, and if some clever filtering has been applied to prevent you from seeing what you want.

    If you are willing to do this, could you attach the report template to a reply?  I'd be happy to take a look to see if I can help you to work out what's (not) happening.

    Kindest regards

    Andy