Hello, I have a standard that has only one peak with which I perform the calibration, but the sample has 2 peaks. How can I calibrate it?, te software is OpenLab CDS C 01.08
Thanks!
Find maintenance, best practice, and troubleshooting articles.
Hello, I have a standard that has only one peak with which I perform the calibration, but the sample has 2 peaks. How can I calibrate it?, te software is OpenLab CDS C 01.08
Thanks!
Hello,
You cannot directly reference the other peaks calibration curve for named peaks. You could manually copy the same response factor in the calibration table or use intelligent reporter to do a custom calculation to reference the other peaks RF or curve details.
Marty