Hi all,
We have two HPLC-MS systems which we are currently setting to standby manually after sequence tables finish running, but would like to have a delayed macro setup to go into standby after a set period of time (i.e. after 15 minutes or 900 seconds). Is there some way to write a shutdown macro so that a timer begins after the sequence table(s) have completed running prior to standing by? This was able to be set in the GUI when I used Waters MassLynx software previously, but it seems to require a macro for ChemStation.
The macro we would like to amend is the default included with ChemStation:
name shutdown
local reply$
print NLSCoreUtil$("ShutdownMsg")
!!!!! Do your own commands here
STANDBY
! switch off FLD lamp
if dev1046FLD = 1
if _offline = 0
reply$ = SendModule$ ("FLD", "LAMP=0")
endif
endif
print NLSCoreUtil$("ShutdownMsg")
endmacro
Thank you.