Module shutdown

In OpenLab CDS V2.5 there is an option to shutdown the detector module after an predefined time in minutes. This option is only available when run queu is idle, is there something like an script or turn around to keep modules running and shutdown the lamp during weekend? 

Parents Reply Children
  • I know that this is getting old now, but did you consider the shutdown method approach.  From the run queue, you can submit a Shutdown run that will sit at the end of the queue until all of your samples have run, and will then turn all modules to standby, including turning off UV lamps.

    I think the only other option is to use the LC Detector Control to set a timeout for detector lamps.  After your long runs at the weekend, this would turn off just the detector lamps, as you need.

    Andy

Was this helpful?