Does anyone know how would I go about obtaining a macro for creating injection logs on MSD Chemstation? As of right now, the only option to print what has run on my 6890s is to print the sequence log(which is much more inconvenient).
Thanks!
Find maintenance, best practice, and troubleshooting articles.
Does anyone know how would I go about obtaining a macro for creating injection logs on MSD Chemstation? As of right now, the only option to print what has run on my 6890s is to print the sequence log(which is much more inconvenient).
Thanks!
Not knowing what revision of MSD chemstation, in G1701EA in data analysis, load a data file and on command line type
macro "injlog.mac",go then hit the execute button
then a log file will pop up.
Hi ron_tackett, thanks for the response!
Product: G1701EA
Revision: E.02.01 1177
When I typed in the command, nothing happend.
Oh ok, I see it now in the data folder. Does it just show up there, or is there a way to make it show up on the screen when you do the command?
This is done on the command line in data analysis after a data is loaded. On command line type
macro "injlog.mac",go hit the execute button.
A log file will pop up. I just tested E.02.01.
Ok, I just figured out my other window was blocking it. This helps a lot, thank you so much!
Ok, I just figured out my other window was blocking it. This helps a lot, thank you so much!