Hi I am using Chemstation Edition software, I am coming across the following error message, since yesterday. If a start a sequence after 2 tube i am getting the below message.
Can anyone help how to come across this.
Kind Regards
Revathy
Find maintenance, best practice, and troubleshooting articles.
Hi I am using Chemstation Edition software, I am coming across the following error message, since yesterday. If a start a sequence after 2 tube i am getting the below message.
Can anyone help how to come across this.
Kind Regards
Revathy
That was very informative. I have Adobe 9. So i need to uninstall and need to install 2017 classic track. Will try and let you know.
Thanks
Kind Regards
Revathy
Hello,
What version of OpenLab ChemStation are you running? You can determine this by going to Help>About
Usually chemmain crashes are triggered when the software is generating a report due to Adobe Reader not being configured correctly. See attached document for recommendations on how to configure Adobe Reader.